Hi Erin! I'm looking at the setup amps function, and it seems to enable amps in this else branch: sodetlib/sodetlib/operations/uxm_setup.py Lines 125 to 131 in 6de534a I can't remember off the top of my head why we do it there, since it looks like it doesn't happen for major version == 4... but I know this is regularly used at site and works relatively well. Can you:
|
Beta Was this translation helpful? Give feedback.
-
Ok thanks! I'll double check the SCSI cables now. The version we are on is: (4,0,7) |
Beta Was this translation helpful? Give feedback.
-
Ok so we remated the SCSI connectors and didn't see a difference. We are now wondering if it has something to do with the S.C.write_ps_en(). We turned the amps off with this function, and when we attempted alter this function, we got a different result, so it seems like we might have disabled our power supplies by accident. @jlashner do you know the correct way to enable the psus, do you think that this could be the source of the issue? |
Beta Was this translation helpful? Give feedback.
-
When we run [ 2024-07-15 18:35:21 ] get_amp_drain_voltage: The power supply for amp 50k1 is off, therefore returning 0.0.
[ 2024-07-15 18:35:21 ] get_amp_drain_voltage: The power supply for amp 50k2 is off, therefore returning 0.0.
[ 2024-07-15 18:35:21 ] get_amp_drain_voltage: The power supply for amp hemt1 is off, therefore returning 0.0.
[ 2024-07-15 18:35:21 ] get_amp_drain_voltage: The power supply for amp hemt2 is off, therefore returning 0.0.
{'50k1_gate_volt': 0.0,
'50k1_drain_volt': 0.0,
'50k1_drain_current': 0.29211306526336983,
'50k1_enable': False,
'50k2_gate_volt': 0.0,
'50k2_drain_volt': 0.0,
'50k2_drain_current': 0.29211306526336983,
'50k2_enable': False,
'hemt1_gate_volt': -2.0048469742400004,
'hemt1_drain_volt': 0.0,
'hemt1_drain_current': -0.05331653329891589,
'hemt1_enable': False,
'hemt2_gate_volt': -1.9999947968000003,
'hemt2_drain_volt': 0.0,
'hemt2_drain_current': -0.07321653329891589,
'hemt2_enable': False} It looks like all of the amplifiers are not enabled. How would we enable them? |
Beta Was this translation helpful? Give feedback.
-
Ran I think it might be an issue with our config file, because the amp currents are way too high. {'50k1_gate_volt': 0.0,
'50k1_drain_volt': 4.000003911132813,
'50k1_drain_current': 55.33919736378285,
'50k1_enable': True,
'50k2_gate_volt': 0.0,
'50k2_drain_volt': 4.000003911132813,
'50k2_drain_current': 55.33919736378285,
'50k2_enable': True,
'hemt1_gate_volt': -2.0048469742400004,
'hemt1_drain_volt': 0.8000042998313902,
'hemt1_drain_current': 33.80703452503872,
'hemt1_enable': True,
'hemt2_gate_volt': -1.9999947968000003,
'hemt2_drain_volt': 0.8000042998313902,
'hemt2_drain_current': 33.787134525038724,
'hemt2_enable': True} |
Beta Was this translation helpful? Give feedback.
-
Replying to request help again! We were able to get our amps to draw the right current with non-smurf power supplies. So this does not seem to be an issue with the amplifiers themselves. |
Beta Was this translation helpful? Give feedback.
-
So I believe the 33 and 55 A readout is not physical, and is a sign that there is a problem with the current readout. We've seen it before in a previous other discussions: We should check if @tpsatt remembers how we solved these, but it seems in at least one of them, hammering the slot fixed the issue so it might be worth just trying that first. Around this time the LAT there was also an issue with the seating of the DSUBs on the cryocard side. When you checked the DC cable, did you check both the SCSI and dsubs? |
Beta Was this translation helpful? Give feedback.
-
Thanks, Jack! Helpful to see this issue was resolved by hammering. Unfortunately, during debugging we have encountered a new issue in that our cryocard does not seem to be turning on. We had disconnected from smurf to check our amplifiers and after powering everything back on, we don't have power to this cryocard anymore. I've tried replacing the SCSI cable. |
Beta Was this translation helpful? Give feedback.
-
Ok so, we've swapped out the SCSI cable, checked and remated the DC cable, and power-cycled the crate and hammered. When we ran setup_amps we initially got the same 55 and 33 mA values, so we hammered again. Now we are getting a new issue: [ 2024-07-16 16:28:59 ] Starting setup_amps
[ 2024-07-16 16:28:59 ] get_amp_drain_voltage: The power supply for amp 50k1 is off, therefore returning 0.0.
cannot connect to smurf_server_s3:AMCc:FpgaTopLevel:AppTop:AppCore:RtmCryoDet:RtmSpiMax:TesBiasDacDataRegCh[26]
[ 2024-07-16 16:29:04 ] Command failed: smurf_server_s3:AMCc:FpgaTopLevel:AppTop:AppCore:RtmCryoDet:RtmSpiMax:TesBiasDacDataRegCh[26]
[ 2024-07-16 16:29:04 ] Retry attempt 1 of 5
cannot connect to smurf_server_s3:AMCc:FpgaTopLevel:AppTop:AppCore:RtmCryoDet:RtmSpiMax:TesBiasDacDataRegCh[26]
[ 2024-07-16 16:29:09 ] Retry attempt 2 of 5
cannot connect to smurf_server_s3:AMCc:FpgaTopLevel:AppTop:AppCore:RtmCryoDet:RtmSpiMax:TesBiasDacDataRegCh[26]
[ 2024-07-16 16:29:14 ] Retry attempt 3 of 5
cannot connect to smurf_server_s3:AMCc:FpgaTopLevel:AppTop:AppCore:RtmCryoDet:RtmSpiMax:TesBiasDacDataRegCh[26]
[ 2024-07-16 16:29:19 ] Retry attempt 4 of 5
cannot connect to smurf_server_s3:AMCc:FpgaTopLevel:AppTop:AppCore:RtmCryoDet:RtmSpiMax:TesBiasDacDataRegCh[26]
[ 2024-07-16 16:29:24 ] Retry attempt 5 of 5
cannot connect to smurf_server_s3:AMCc:FpgaTopLevel:AppTop:AppCore:RtmCryoDet:RtmSpiMax:TesBiasDacDataRegCh[26] And the error was that epics couldn't connect. We tried hammering again but got this same error. |
Beta Was this translation helpful? Give feedback.
-
Restarted jupyter docker and jupyter notebook. Seemed to solve all of these problems. Thanks for the help @jlashner! Was able to use some old github discussions to continue debugging. |
Beta Was this translation helpful? Give feedback.
-
We are running into an issue when we try to set up amps. We are getting an error about a power supply being off. I am wondering if we have an issue with a config file setting or something, because I think we are using the right script (setup_amps).