Le Petit Prince sub-EN100, post-fMRIPrep denoised data look odd #899

Answered by handwerkerd
tsalo asked this question in Q&A
Discussion options

You must be logged in to vote

The highest variance components have a very large spike at the beginning of the scans. That is usually a sign that the non-steady state volumes weren't removed. Siemens removes those by default, but GE keeps them. Looking at the openneuro description of these data, some participants were acquired on a Siemens scanner and others on a GE scanner. Any chance this run was GE scanner without steady state volumes removed?

Replies: 1 comment 5 replies

Comment options

You must be logged in to vote
5 replies
@tsalo
Comment options

tsalo Nov 17, 2022
Maintainer Author

@tsalo
Comment options

tsalo Nov 18, 2022
Maintainer Author

@handwerkerd
Comment options

@tsalo
Comment options

tsalo Nov 18, 2022
Maintainer Author

@tsalo
Comment options

tsalo Nov 19, 2022
Maintainer Author

Answer selected by tsalo